Linear Gage LGD-Absolute,Standard Dimensions, Robust Series 575
Series 575
Resolution: 10μm
  • Air drive unit
    For 10mm range models: No.02ADE230
    For 25mm range models: No.02ADE250
    For 50mm range models: No.02ADE270
    * Required air pressure: 0.2 to 0.4MPa
    * Spindle extends when air is supplied.
  • Rubber boot (spare)
    For 10mm range models: No.238772
    For 25mm range models: No.962504
    For 50mm range models: No.962505
  • Thrust stem set
    For 10mm range models: No.02ADB680
    Thrust stem: No.02ADB681
    Clamp nut: No.02ADB682
    For 25/50mm range models: No.02ADN370
    Thrust stem: No.02ADN371
    Clamp nut: No.02ADB692
    * External dimensions are described in the dimensional drawing of the product.
    * Thrust stem set is a combination of thrust stem and a clamp nut. A special spanner is required for tightening. If using multiple gages, a thrust stem set for each gage and one special spanner are required.
  • Spanner Wrench
    For 10mm range models: No.02ADB683
    For 25/50mm range models: No.02ADB693
    Digimatic Power Supply Unit: 965370
    SPC cable extension adapter: 02ADF640
    Extension cable (0.5m): 02ADD950
    Extension cable (1m): 936937
    Extension cable (2m): 965014
    *when connecting an extension cable, an SPC cable extension adapter is required (02ADF640)
    Applicable Counters
    542-007A EC-101D Counter, 120V
    64PKA132 EG-101D
    64PKA135 EB-11D
    542-072A EH-102D
    542-064 EV-16D COUNTER
*1: The last number of the Code No. represents special cable length. (meters)
*2: The absolute reference point is near the lowest rest point at shipment.
*3: Mass including 2m cable.
*4: IP code is a standard which classifies and rates the degree of protection    provided against the intrusion of solid objects and water. This may not be     applicable depending on the type of liquid.
Metric
Order No*1575-326575-327575-328
Range 10mm25mm50mm
Resolution.0005"/10µm.0005"/10µm.0005"/10µm
Accuracy 20µm20µm30µm
Quantizing error±1 count±1 count±1 count
Contact point upward1.0N or less4.0N or less4.9N or less
Contact point horizonal1.1N or less4.3N or less5.3N or less
Contact point downward1.2N or less4.6N or less5.7N or less
Stem Diaø8mmø15mmø15mm
Position detection methodABSOLUTE electrostatic capacitance-type linear encoderABSOLUTE electrostatic capacitance-type linear encoderABSOLUTE electrostatic capacitance-type linear encoder
Response speedUnlimited (not applicable to scanning measurement)Unlimited (not applicable to scanning measurement)Unlimited (not applicable to scanning measurement)
OutputDigimatic outputDigimatic outputDigimatic output
External inputReference-setting signal (Absolute reference position*2) can be changed externally.Reference-setting signal (Absolute reference position*2) can be changed externally.Reference-setting signal (Absolute reference position*2) can be changed externally.
Mass *3Aprox.260gAprox.300gAprox.400g
Contact pointø3mm carbide-tipped (fixing screw: M2.5 (P=0.45)×5), standard contact point No.901312ø3mm carbide-tipped (fixing screw: M2.5 (P=0.45)×5), standard contact point No.901312ø3mm carbide-tipped (fixing screw: M2.5 (P=0.45)×5), standard contact point No.901312
Stem dia.ø8ø15ø15
Bearing typeLinear ball bearingLinear ball bearingLinear ball bearing
Dust/water resistance*4Equivalent to IP66 (only gage head)Equivalent to IP66 (only gage head)Equivalent to IP66 (only gage head)
Output cable length
(directly extended from the main unit)
2m,3m,5m,7m2m,3m,5m,7m2m,3m,5m,7m
Operating temperature (humidity) range0 to 40°C (RH 20 to 80%, no condensation0 to 40°C (RH 20 to 80%, no condensation0 to 40°C (RH 20 to 80%, no condensation
Storage temperature(humidity) range–10 to 60°C (RH 20 to 80%, no condensation–10 to 60°C (RH 20 to 80%, no condensation–10 to 60°C (RH 20 to 80%, no condensation